Vol.:(0123456789) 1 3 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 https://doi.org/10.1007/s10334-019-00806-7 RESEARCH ARTICLE Comparison of multi‑delay FAIR and pCASL labeling approaches for renal perfusion quantification at 3T MRI Anita A. Harteveld 1  · Anneloes de Boer 1  · Suzanne Lisa Franklin 1,2  · Tim Leiner 1  · Marijn van Stralen 1  · Clemens Bos 1 Received: 21 August 2019 / Revised: 29 October 2019 / Accepted: 19 November 2019 / Published online: 6 December 2019 © The Author(s) 2019 Abstract Objective To compare the most commonly used labeling approaches, flow-sensitive alternating inversion recovery (FAIR) and pseudocontinuous arterial spin labeling (pCASL), for renal perfusion measurement using arterial spin labeling (ASL) MRI. Methods Multi-delay FAIR and pCASL were performed in 16 middle-aged healthy volunteers on two different occasions at 3T. Relative perfusion-weighted signal (PWS), temporal SNR (tSNR), renal blood flow (RBF), and arterial transit time (ATT) were calculated for the cortex and medulla in both kidneys. Bland–Altman plots, intra-class correlation coefficient, and within-subject coefficient of variation were used to assess reliability and agreement between measurements. Results For the first visit, RBF was 362 ± 57 and 140 ± 47 mL/min/100 g, and ATT was 0.47 ± 0.13 and 0.70 ± 0.10 s in cortex and medulla, respectively, using FAIR; RBF was 201 ± 72 and 84 ± 27 mL/min/100 g, and ATT was 0.71 ± 0.25 and 0.86 ± 0.12 s in cortex and medulla, respectively, using pCASL. For both labeling approaches, RBF and ATT values were not significantly different between visits. Overall, FAIR showed higher PWS and tSNR. Moreover, repeatability of perfusion parameters was better using FAIR. Discussion This study showed that compared to (balanced) pCASL, FAIR perfusion values were significantly higher and more comparable between visits. Keywords Arterial spin labeling (ASL) · Arterial transit time · Magnetic resonance imaging · Kidney · Renal blood flow Introduction Renal perfusion is a valuable physiological parameter for assessing kidney function and identifying pathology [1]. In recent years, arterial spin labeling (ASL) magnetic reso- nance imaging (MRI) has been emerging as a method for measurement of renal perfusion [2] that does not warrant administration of contrast agent [35] or the involvement of lengthy invasive clearance measurements [6, 7]. An addi- tional advantage of ASL is that it enables measurement of local perfusion, in contrast to clearance techniques which only assess renal blood flow of both kidneys combined. In ASL, images are acquired with (label) and without (control) inverting/saturating magnetization of arterial blood flowing into the tissue. The signal intensity present in the subtracted label–control images is proportional to perfusion of the tis- sue [8]. Renal perfusion is a relatively new application area of ASL-MRI. In the white paper for brain ASL [9], pseudo- continuous ASL (pCASL) is the recommended labeling approach. However, the most commonly used labeling approach for renal ASL thus far has been flow-sensitive alternating inversion-recovery pulsed ASL (FAIR) [2]. In general, FAIR has the advantage of higher labeling effi- ciency and lower specific absorption rate, whereas pCASL Electronic supplementary material The online version of this article (https://doi.org/10.1007/s10334-019-00806-7) contains supplementary material, which is available to authorized users. * Anita A. Harteveld [email protected] 1 Department of Radiology, University Medical Center Utrecht, Utrecht University, Postbox 85500, 3508 GA Utrecht, The Netherlands 2 Department of Radiology, C.J. Gorter Center for High Field MRI, Leiden University Medical Center, Leiden, The Netherlands

Transcript of C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in...

Page 1: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

Vol.:(0123456789)1 3

Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 https://doi.org/10.1007/s10334-019-00806-7

RESEARCH ARTICLE

Comparison of multi‑delay FAIR and pCASL labeling approaches for renal perfusion quantification at 3T MRI

Anita A. Harteveld1  · Anneloes de Boer1 · Suzanne Lisa Franklin1,2 · Tim Leiner1 · Marijn van Stralen1 · Clemens Bos1

Received: 21 August 2019 / Revised: 29 October 2019 / Accepted: 19 November 2019 / Published online: 6 December 2019 © The Author(s) 2019

AbstractObjective To compare the most commonly used labeling approaches, flow-sensitive alternating inversion recovery (FAIR) and pseudocontinuous arterial spin labeling (pCASL), for renal perfusion measurement using arterial spin labeling (ASL) MRI.Methods Multi-delay FAIR and pCASL were performed in 16 middle-aged healthy volunteers on two different occasions at 3T. Relative perfusion-weighted signal (PWS), temporal SNR (tSNR), renal blood flow (RBF), and arterial transit time (ATT) were calculated for the cortex and medulla in both kidneys. Bland–Altman plots, intra-class correlation coefficient, and within-subject coefficient of variation were used to assess reliability and agreement between measurements.Results For the first visit, RBF was 362 ± 57 and 140 ± 47 mL/min/100 g, and ATT was 0.47 ± 0.13 and 0.70 ± 0.10 s in cortex and medulla, respectively, using FAIR; RBF was 201 ± 72 and 84 ± 27 mL/min/100 g, and ATT was 0.71 ± 0.25 and 0.86 ± 0.12 s in cortex and medulla, respectively, using pCASL. For both labeling approaches, RBF and ATT values were not significantly different between visits. Overall, FAIR showed higher PWS and tSNR. Moreover, repeatability of perfusion parameters was better using FAIR.Discussion This study showed that compared to (balanced) pCASL, FAIR perfusion values were significantly higher and more comparable between visits.

Keywords Arterial spin labeling (ASL) · Arterial transit time · Magnetic resonance imaging · Kidney · Renal blood flow

Introduction

Renal perfusion is a valuable physiological parameter for assessing kidney function and identifying pathology [1]. In recent years, arterial spin labeling (ASL) magnetic reso-nance imaging (MRI) has been emerging as a method for measurement of renal perfusion [2] that does not warrant

administration of contrast agent [3–5] or the involvement of lengthy invasive clearance measurements [6, 7]. An addi-tional advantage of ASL is that it enables measurement of local perfusion, in contrast to clearance techniques which only assess renal blood flow of both kidneys combined. In ASL, images are acquired with (label) and without (control) inverting/saturating magnetization of arterial blood flowing into the tissue. The signal intensity present in the subtracted label–control images is proportional to perfusion of the tis-sue [8].

Renal perfusion is a relatively new application area of ASL-MRI. In the white paper for brain ASL [9], pseudo-continuous ASL (pCASL) is the recommended labeling approach. However, the most commonly used labeling approach for renal ASL thus far has been flow-sensitive alternating inversion-recovery pulsed ASL (FAIR) [2]. In general, FAIR has the advantage of higher labeling effi-ciency and lower specific absorption rate, whereas pCASL

Electronic supplementary material The online version of this article (https ://doi.org/10.1007/s1033 4-019-00806 -7) contains supplementary material, which is available to authorized users.

* Anita A. Harteveld [email protected]

1 Department of Radiology, University Medical Center Utrecht, Utrecht University, Postbox 85500, 3508 GA Utrecht, The Netherlands

2 Department of Radiology, C.J. Gorter Center for High Field MRI, Leiden University Medical Center, Leiden, The Netherlands

Page 2: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

can achieve a higher intrinsic signal-to-noise ratio (SNR) [9, 10].

ASL-MRI in the abdomen brings new challenges with respect to the brain, like respiratory motion, complex vas-culature, and increased magnetic field inhomogeneities due to the proximity of air in the lungs and in the digestive tract. Thus far, no studies have been performed directly comparing different labeling approaches for renal ASL, and therefore, the effect on the obtained perfusion signal remains unclear. The aim of this study was to compare ASL-MRI of the kid-neys with the two most commonly used labeling approaches FAIR and pCASL, to obtain a better insight into the reliabil-ity and repeatability of each method and to identify the most efficient method to perform renal perfusion measurements.

Methods

Study population

Between March and November 2018, volunteers aged > 40 years, without history of renal disease or contraindica-tions for MR imaging, were included. This prospective study was approved by the local institutional review board and all subjects provided written informed consent. The scans used in this study were acquired as part of the multiparametric repeatability ReMaRK study (Repeatability of functional Magnetic Resonance imaging of the Kidneys).

MR imaging

Imaging was performed on a 3T MR system (Ingenia, Philips, Best, The Netherlands; release 5.3.1) using a body coil for transmission and a 28-element phased array coil for reception. All subjects were scanned twice with an inter-val of 1 week between acquisitions (median 7 days; range 4–14 days). To control gastrointestinal physiological condi-tions, both visits were scheduled at the same time of the day (afternoon), and subjects were asked to drink 2 L of non-alcoholic liquids and to avoid salt and protein rich meals prior to the MRI examination. Prior to the first-scan session, blood was sampled from each subject to determine kidney function by calculating the eGFR with the CKD-EPI formula (Chronic Kidney Disease EPIdemiology collaboration [11]) using the measured creatinine.

Scan protocol

The scan protocol consisted of both FAIR and pCASL per-fusion imaging, both with varying delay times between labeling and readout, and auxiliary sequences to estimate equilibrium magnetization (M0) and T1 for perfusion quantification.

FAIR labeling was implemented as previously described [12]. In this method, label and control conditions are achieved by alternating between a selective and non-selec-tive inversion slab. An adiabatic FOCI (frequency offset-cor-rected inversion) pulse was used for the selective inversion [13, 14]. Pre-saturation using WET [15] (water suppression enhanced through T1 effects) consisting of four pulses and post-saturation using a single 90° pulse were applied to the imaging region directly before and after each selective or non-selective inversion pulse, respectively, to minimize perfusion-weighted signal differences caused by inversion efficiency differences between both inversion slabs. The selective inversion slab was aligned with the imaging stack and was 10 mm wider (thickness 54 mm) than this stack (thickness 34 mm). To define the temporal bolus width of the labeled spins, a QUIPSS II scheme was applied at a spe-cific delay time (TI1) after the inversion pulse (five saturation pulses timed equidistantly within 100 ms after TI1) that was placed anterior of the imaging stack with a gap of 10 mm with the aim to cover the feeding arteries, for the kidneys specifically the descending aorta [16, 17]. The thickness of this saturation slab was 120 mm. FAIR data were acquired with four different times-to-inversion (TI; 0.8, 1.4, 2.0, and 2.6 s) and QUIPSS II at TI1 of 1.2 s.

pCASL labeling was performed using the balanced ver-sion [18], as implemented by the vendor. In this method, flowing spins are inverted by a long train of short, repeated RF pulses (Hanning-shaped, duration 480  μs, spacing 1210 μs, average B1 1.5 μT) in combination with a switching selection gradient (average strength 0.36 mT/m) to achieve the label condition. WET water suppression was applied to the imaging region directly before the labeling, to elimi-nate residual magnetization modulation from the previous acquisition. pCASL data were acquired with four different post-labeling delays (PLD; 0.5, 1.0, 1.5, and 2.0 s) and a label duration of 1.5 s.

Each delay time was obtained in a separate acquisition that consisted of ten label–control pair repetitions. The M0 scan, essentially the FAIR/pCASL scan without labeling or suppression pulses, was acquired four times to improve the SNR. The T1 map was acquired using a cycled multislice inversion-recovery sequence [19] with 11 inversion times (range 55–2035 ms). All scans were performed with the same single-shot gradient echo EPI 2D multislice readout. Detailed scan parameters are provided in Table 1.

A localizer scan was acquired to enable correct planning of all subsequent scans. The image readout of all scans was planned coronal oblique along the long axis of the kidneys to minimize through plane motion of the kidneys; the center slice had identical geometry for all scans. For FAIR scans, care was taken to exclude the feeding arteries of the kid-neys, especially the descending aorta, from the selective inversion slab and to include them within the QUIPSS II

Page 3: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

83Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

slab. For pCASL scans, the same procedure was applied during each scan session to plan the labeling slab. The labe-ling slab was placed perpendicular to the descending aorta, approximately 10–11 cm above the center of the kidneys. Care was taken that the labeling slab was placed above the kidneys to prevent the kidneys from sliding into the labeling slab during breathing, but below the diaphragm to minimize susceptibility artifacts from the air–tissue interface at the lungs. The planning of both labeling approaches is shown in Supplementary Figure S1. B0 shimming was applied to the imaging stack and the labeling slab independently. B1 shimming was applied to a volume shim box that covered the entire field-of-view. All acquisitions were performed with paced breathing. Subjects were instructed before scanning and coached during scanning to synchronize breathing to the TR of 6500 ms of the sequence and to briefly hold their breath in expiration during the image readout [20]. Labeling was timed directly after expiration. A bellows was placed on the subject’s upper abdomen to monitor respiration and the compliance with the breathing instructions.

Image processing

Image processing and analysis were performed using custom scripts in MeVisLab (version 2.8.2; MeVis Medical Solu-tions AG, Bremen, Germany). Processing was performed per subject on images obtained during the same visit, separately for each kidney. After a wide crop around each kidney was made, slice-wise motion correction was performed using a groupwise image registration method [21] implemented in Elastix [22] to compensate for misalignments due to (res-piratory) motion. ASL, M0, and T1 images were registered simultaneously. To this end, 3D slice stacks were created for each slice of the cropped image, with the non-subtracted ASL (multi-delay FAIR and pCASL), M0 and T1 slices in the third dimension, with a total of 175 images per registra-tion; see Fig. 1. The groupwise registration method does not require the choice of a reference image, thus avoiding registration bias, and the method is robust against intensity changes between the images.

After image registration, ASL label and control images were pair-wise subtracted (ΔM). Next, for each ASL scan (delay time), outlier rejection was performed by excluding subtraction images containing > 20% voxels (within the

Table 1 Scan parameters

ASL arterial spin labeling, TR repetition time, TE echo time, EPI echo planar imaging, SENSE sensitivity encoding, FOV field-of-view, SPIR spectral pre-saturation with inversion recovery, FH feet–head, AP ante-rior–posteriora Spatial saturation slabs superior and inferior to the image volume to suppress undesired signal aliasingb Label–control pairs per delay timec QUIPSS II saturation pulses were set at TI1 = 1.2 s; delay time 0.8 s was, therefore, obtained without these saturation pulses

Parameters ASL M0 T1 map

TR/TE (ms) 6500/22 6500/22 6500/22EPI factor 65 65 65Flip angle (°) 90 90 90SENSE 1.5 (FH direction) 1.5 (FH direction) 1.5 (FH direction)FOV (mm3) 244 × 244 × 34 244 × 244 × 48 244 × 244 × 76Acquired voxel size (mm2) 3 × 3 3 × 3 3 × 3Slice thickness (mm) 6 6 6Slice gap (mm) 1 1 1No. of slices FAIR: 5

pCASL: 77 11

Phase encoding direction FH FH FHFold-over suppression Saturation slabsa Saturation slabsa Saturation slabsa

Fat suppression SPIR SPIR SPIRSlice orientation Coronal Coronal CoronalSlice scan order AP AP APNo. of repetitions 10b 4 N/ADelay time (s) FAIRc: 0.8, 1.4, 2.0, 2.6

pCASL: 0.5, 1.0, 1.5, 2.0N/A N/A

Inversion time (ms) N/A N/A 55–2035; steps 198Total acquisition time (min:s) 02:23 00:32.5 01:11.5

Page 4: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

84 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

kidney region) with a value of more than ± 2 SD from the mean voxel value over all repetitions. The remaining sub-traction images were averaged per delay time to obtain the perfusion-weighted images ( ΔM ). Voxel-wise T1 relaxation times were calculated by a nonlinear least-squares fit of the 11 inversion-recovery magnitude images to a monoexponen-tial recovery function. The M0 images were averaged over the four repetitions. The whole kidney region that was used for the outlier rejection was manually drawn along the vis-ible kidney contour on the averaged M0 images after image registration.

Voxel-wise perfusion quantification was performed using Buxton’s general kinetic model for pulsed and continuous labeling ASL signal to estimate renal blood flow (RBF) and arterial transit time (ATT) [23]. The measured ΔM , M0, and T1,tissue values were used as input for the two-compartment model together with assumptions for T1,blood of 1.65 s [24], tissue–blood partition coefficient of 0.9 mL/g [25], and labe-ling efficiency of 95% for FAIR and 85% for pCASL. In the model fit, the exact delay time of each slice was used by taking into account the delay between slices of 65 ms.

The number of slices for the FAIR acquisition was limited by the available space between the slice-selective inversion slab and the aorta. Therefore, only five slices were pro-cessed, analyzed, and compared for FAIR and pCASL.

Image analysis

For quantitative comparison of both labeling approaches, the following metrics were used: (1) relative perfusion-weighted signal (PWS), (2) temporal SNR (tSNR), (3) RBF, and (4) ATT. PWS was calculated for each delay time as ΔM/M0 × 100%. tSNR was defined as �ΔM∕σΔM where �ΔM is the temporal voxel-wise mean and σΔM is the temporal voxel-wise standard deviation of the subtraction (ΔM) images. tSNR was also calculated for each delay time separately. The tSNR was used as a metric for consistency of the perfusion

signal over repetitions. For separate analysis of the cortex and medulla, semi-automatic segmentation based on the intensity histogram of the T1 map was performed by a single observer to determine thresholds separating both regions. Because of the groupwise registration of FAIR and pCASL, the same segmentations were used for both analyses. When necessary, segmentations were manually adjusted with in-plane affine transformations.

Statistical analysis

For the quantitative metrics (PWS, tSNR, RBF, and ATT), mean and SD values were calculated over all voxels inside the segmented kidney regions (cortex and medulla) for each subject. Inter-visit reliability and agreement between meas-urements (i.e., repeatability) of RBF and ATT were evalu-ated using the intra-class correlation coefficient (ICC), within-subject coefficient of variation (CVws), and Bland–Altman analysis. The CVws was calculated as (SD/mean) × 100% with SD =

(∑

(x1− x

2)2)∕2n , where x1 and

x2 are the measurement values at both visits in the same subject, n is the number of subjects, and the mean is the average over all measurements. Differences in RBF and ATT between visits, between labeling approaches, and between left and right kidney, were tested with a Wilcoxon matched-pair signed-rank test. Spearman’s rho was used to determine association of perfusion values (RBF and ATT) between both kidneys, and between RBF and eGFR. Statistical analy-ses were performed using R [26] version 3.6.1. A P value of < 0.05 was considered to be statistically significant.

Fig. 1 Example images of the right kidney before and after image registration. Improved alignment can be observed for the images after motion correc-tion. Images visualize the same intersection (yellow line in the coronal image) over all M0, ASL, and T1 images

Page 5: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

85Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

Results

Study population

Sixteen middle-aged volunteers (eight male; age 51 ± 10 years) were included. All subjects had an eGFR in the normal range (86 ± 15 mL/min/1.73m2). Five data sets from four subjects were excluded from analysis due to: (1) absence of one of the labeling methods (one subject, both visits), (2) presence of severe motion differences between FAIR and pCASL (two subjects, second visit), or (3) MRI exam was not performed (one subject, second visit). This resulted in 15 complete data sets (including both FAIR and pCASL) from the first visit and 12 available complete data sets in the same subjects from the second visit that were used for analysis.

Image processing and analysis

Improved image alignment after motion correction was observed in all subjects (Fig. 1). Outlier rejection was per-formed in 18/27 FAIR and 19/27 pCASL data sets, with a maximum of two excluded label–control pairs per delay time (Fig. 2). Manual adjustment of segmentations was necessary for 7/54 data sets of the left (n = 4) and right (n = 3) kidney. Example images after image processing and analysis of the obtained scans are shown in Fig. 3, demonstrating a clear PWS and RBF contrast between renal cortex and medulla for both labeling techniques.

Quantitative metrics

PWS and tSNR

PWS and tSNR values obtained with both labeling approaches are shown in Fig.  4. For both labeling approaches, cortical PWS showed an initial increase from the first to the second delay time and a decrease after the second delay time. For pCASL, the inter-subject variability in the PWS was much higher than for FAIR. The medullary PWS only decreased with increasing delay time. The tSNR showed similar behavior as the PWS, consistent with similar noise levels for each delay time. Although evaluated at dif-ferent delay times, cortical and medullary PWS and tSNR were in general higher with FAIR, at least 37%.

RBF

Mean cortical RBF was 362 ± 57  mL/min/100  g with FAIR and 201 ± 72 mL/min/100 g with pCASL at visit 1 (P < 0.001). Mean medullary RBF was 140 ± 47 mL/min/100  g with FAIR and 84 ± 27  mL/min/100  g with

pCASL at visit 1 (P < 0.001). For both FAIR and pCASL, RBF values were significantly different neither between vis-its (P ≥ 0.34; Table 2), nor between left and right kidney of a subject (P = 0.93 and P = 0.52 in cortex; P = 0.42 and P = 0.98 in medulla; for FAIR and pCASL, respectively). There was a significant correlation between the measured RBF values in the left and right kidney of each subject, both with FAIR (r = 0.83, P < 0.001 in cortex and r = 0.86, P < 0.001 in medulla), and pCASL (r = 0.94, P < 0.001 in cortex and r = 0.58, P = 0.025 in medulla) (Fig. 5). No cor-relation was observed between RBF and eGFR for either approach (r ≤ 0.45, P > 0.05; data not shown).

Bland–Altman plots comparing RBF measurements between visits for both labeling approaches are shown in Figs. 6 and 7 for cortex and medulla, respectively. A better inter-visit agreement was observed with FAIR in both the cortex and medulla. The ICC for FAIR was moderate for cortex and good for medulla, and the CVws was relatively low for both kidney regions. The ICC for pCASL was poor, and the CVws was high for both kidney regions (Table 2). On average, RBF values measured using FAIR were 161 mL/min/100 g higher for cortex and 56 mL/min/100 g higher for medulla, than when measured using pCASL (Fig. 8).

ATT

Mean cortical ATT was 0.47 ± 0.13  s with FAIR and 0.71 ± 0.25 s with pCASL at visit 1. Mean medullary ATT was 0.70 ± 0.10 s with FAIR and 0.86 ± 0.12 s with pCASL at visit 1. For both labeling approaches, ATT values were not significantly different between visits (P ≥ 0.18; Table 2), and neither between left and right kidneys of each subject for pCASL (P = 0.49 in cortex; P = 0.45 in medulla). How-ever, for FAIR the ATT was significantly different between left and right kidneys of each subject (0.50 ± 0.13 and 0.45 ± 0.14 s, P = 0.010 in cortex for left and right kidneys, respectively; 0.74 ± 0.10 and 0.67 ± 0.12 s, P = 0.022 in medulla for left and right kidneys, respectively). There was a significant correlation between the measured ATT val-ues in the left and right kidneys of each subject, both with FAIR (r = 0.86, P < 0.001 in cortex and r = 0.64, P = 0.011 in medulla), and pCASL (r = 0.95, P < 0.001 in cortex and r = 0.80, P < 0.001 in medulla) (Fig. 5).

Bland–Altman plots comparing ATT measurements between visits for both labeling approaches are shown in Figs. 6 and 7 for cortex and medulla, respectively. A bet-ter inter-visit agreement was observed with FAIR in the cortex, whilst agreement was comparable for both labeling approaches in the medulla. The ICC for FAIR was good for cortex and poor for medulla, and the CVws was relatively low for both kidney regions. The ICC for pCASL was moderate for both kidney regions, and the CVws was high for cortex and relatively low for medulla (Table 2).

Page 6: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

86 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

Discussion

The current study focused on comparing multi-delay ASL measurements with the two labeling approaches most com-monly used in the kidney, FAIR and pCASL. The goal was

to obtain a better insight into the reliability and repeat-ability of each method and to identify the most efficient method to perform multi-delay renal perfusion quantifica-tion at 3T. The study showed that measured renal perfu-sion values depend on the labeling approach. Perfusion values were significantly higher for FAIR than for pCASL,

Fig. 2 Example of outlier rejection to exclude subtraction images. Images show masked subtraction images of the left kidney with pCASL labeling before outlier rejection. Subtraction images contain-ing > 20% voxels with a value of more than ± 2 SD from the mean voxel value over all repetitions were rejected. The label–control pairs

(repetitions) that were removed after outlier rejection are highlighted in red. The color bar indicates PWS [%]. A similar example with FAIR labeling in the same subject during the same visit is shown in the Supplementary Figure S2

Page 7: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

87Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

and showed substantially larger variability between visits for pCASL compared to FAIR.

Previous renal ASL studies at 3T in healthy volunteers have reported mean cortical RBF in the study population ranging from 199–399 mL/min/100 g and 138–296 mL/min/100 g for FAIR and pCASL, respectively [2]. Studies

using multiple delay times reported mean values ranging from 151–309 ml/min/100 g and 117–215 mL/min/100 g for FAIR and pCASL, respectively [27]. These values high-light a wide range between studies that used the same labe-ling approach, as well as indicate a tendency of FAIR to measure higher RBF values, as found in our results. The

Fig. 3 Example of processed images of the right kidney from a 41-year-old female healthy volunteer acquired at the first visit. After alignment of all scans, the multi-delay perfusion-weighted images ( ΔM ), T1 map (T1-relaxation time in s), and M0 scan were used to

quantify RBF (in mL/min/100  g) and ATT (in s). Both labeling approaches were obtained with four different delay times (FAIR: 0.8, 1.4, 2.0, 2.6 s; pCASL: 0.5, 1.0, 1.5, 2.0 s). ATT arterial transit time, RBF renal blood flow

Fig. 4 PWS (a) and tSNR (b) values obtained with FAIR and pCASL labeling approaches in the cortex (black) and medulla (red) averaged over 15 subjects at visit 1 using four different delay times. Error bars represent the standard deviation. PWS perfusion-weighted signal, tSNR temporal signal-to-noise ratio

Page 8: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

88 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

Table 2 RBF and ATT values obtained with multi-delay FAIR and pCASL labeling approaches in the cortex and medulla averaged over all sub-jects at two different visits (mean ± SD)

ATT arterial transit time, CI confidence interval, CVws within-subject coefficient of variation, ICC intra-class correlation coefficient, RBF renal blood flow, SD standard deviationa Based on 15 subjectsb Based on 12 subjectsc Group differences between visits were tested with Wilcoxon matched-pair signed-rank testd Two-way model, absolute agreement, single measures

Visit 1a Visit 2b P valuec CVws [%] ICCd (95%-CI)

FAIR RBF (mL/min/100 g)  Cortex 362 (57) 389 (55) 0.42 9.9 0.51 (− 0.058 to 0.83)  Medulla 140 (47) 135 (32) 0.34 13.8 0.80 (0.46–0.94)

 ATT (s)  Cortex 0.47 (0.13) 0.50 (0.13) 0.18 10.7 0.83 (0.53–0.95)  Medulla 0.70 (0.10) 0.71 (0.06) 0.23 8.1 0.40 (− 0.18 to 0.78)

pCASL RBF (mL/min/100 g)  Cortex 201 (72) 207 (64) 0.57 33.9 − 0.38 (− 0.86 to 0.27)  Medulla 84 (27) 85 (41) 0.85 30.9 0.41 (− 0.23 to 0.79)

 ATT (s)  Cortex 0.71 (0.25) 0.68 (0.23) 0.79 19.4 0.60 (0.069–0.86)  Medulla 0.86 (0.12) 0.88 (0.12) 0.34 9.8 0.53 (− 0.008 to 0.83)

Fig. 5 Scatterplots showing RBF and ATT values in the left and right kidney obtained with FAIR and pCASL labeling approaches in the cortex (black) and medulla (red) from 15 subjects at visit 1. ATT arterial transit time, RBF renal blood flow

Page 9: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

89Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

wide variation of perfusion values makes it difficult to com-pare results between studies, even when the same labeling approach was used. Apart from differences in the used scan-ner hardware and image readout [28], choices of parameters, such as labeling efficiency, blood-tissue partition coefficient, used in the kinetic model for perfusion quantification largely influence the obtained values. In this study, RBF values obtained with pCASL may be underestimated due to a lower labeling efficiency then assumed in the quantification. For renal application, the pCASL labeling efficiency is most likely influenced by a combination of several factors which will be discussed in more detail below.

In this study at 3T, repeatability was better for FAIR compared to pCASL. For some pCASL data sets (3 out of 27), the averaged perfusion-weighted images showed very low signal corresponding to cortical RBF values < 100 mL/min/100 g. Compared with RBF values obtained at the other visit or with FAIR labeling in the same subject, it seems that these pCASL measurements failed. In the Supplementary Figures S3, S4 and Table S1, pCASL results are presented

after exclusion of these data sets. Exclusion of these data sets made the repeatability of pCASL much more compa-rable with FAIR. After elimination, FAIR still gave RBF values that were on average 147 mL/min/100 g higher for cortex and 54 mL/min/100 g higher for medulla than when measured using pCASL.

Despite hydration instructions and planning both visits at the same time of the day, intra-subject variation of per-fusion values between visits was observed for both labe-ling approaches. Physiological variation of blood flow in the aorta [29, 30] might play a role here in the amount of labeled blood that is created and delivered to the kidneys. Although there was intra-subject variability between visits, the diagnostic value of perfusion imaging, showing regional differences within or differences between kidneys, might still be unaffected. Moreover, studies comparing patients with impaired kidney function and healthy controls found rela-tively large perfusion differences of the average cortical RBF between both groups, ranging from 66 to 202 mL/min/100 g [31, 32].

Fig. 6 Bland–Altman plots comparing cortical RBF and ATT measurements between visits for both labeling approaches. Data points represent 24 kidneys from 12 subjects available at both visits. Blue and red dotted lines cor-respond to mean difference and limits of agreement, respec-tively. ATT arterial transit time, RBF renal blood flow

Page 10: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

90 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

Fig. 7 Bland–Altman plots comparing medullary RBF and ATT measurements between visits for both labe-ling approaches. Data points represent 24 kidneys from 12 subjects available at both visits. Blue and red dotted lines cor-respond to mean difference and limits of agreement, respec-tively. ATT arterial transit time, RBF renal blood flow

Fig. 8 Bland–Altman plots comparing RBF measurements between labeling approaches for cortex and medulla. Data points represent 30 kidneys from 15 subjects at the first visit only. Blue and red dotted lines cor-respond to mean difference and limits of agreement, respec-tively. RBF renal blood flow

Page 11: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

91Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

The number of studies reporting reliability and repeat-ability of renal ASL is limited, especially at 3T [2]. Previous studies using FAIR at 3T in healthy volunteers [33, 34] have reported inter-visit ICCs of 0.80 and 0.85, and CVs of 9.2 and 9.3% for cortical RBF. To our knowledge, no studies have been published yet reporting inter-visit reliability and repeatability at 3T for pCASL. However, there is one study that has investigated intra-visit reliability and repeatability of pCASL at 3T [35] reporting an ICC of 0.93, and CV of 14.4% for cortical RBF. Reliability and repeatability of ATT measurements have been investigated even less. One study reported intra-visit an ICC of 0.32, and CV of 33.6% for cortical ATT with pCASL at 3T [35].

The ATT is dependent on the used measurement method, and was, therefore, not directly compared between both labeling approaches. Previous studies using multiple delay times reported average cortical ATT values of 0.11–0.30 s for FAIR and 0.96–1.23 s for pCASL [27]. The obtained ATT is very much determined by the timing of the (first) delay times. Based on preliminary measurements in healthy volunteers, we chose four equidistantly spaced delay times covering the perfusion signal over time both before and after the expected perfusion signal peak in the cortex for most subjects, to enable proper fitting of the perfusion model for quantification. The perfusion signal curves averaged over all subjects (Fig. 4) indicate that the delay times were cor-rectly placed to capture the signal peak, for both labeling approaches. To improve the accuracy of the ATT estima-tion, more measurements with short delay times could be added. In general, accuracy of perfusion estimation increases with the number of sampled delay times [36]. However, for clinical applicability of renal ASL, total acquisition time is restricted, limiting the number of delay times (and label–control repetitions) that can be obtained.

Measuring medullary perfusion is challenging. The medulla is much less perfused compared to the cortex (only ~ 10% of blood flowing into the kidneys flows through the medulla [37]), and the transit time of labeled blood is much longer, which results in more T1 decay before entering the medulla and lower tSNR, as was observed in this study. It is thus not surprising that medullary perfusion showed lower repeatability than cortical perfusion. Thus far, not many studies have reported medullary RBF and ATT values, and mainly focused on measuring cortical perfusion.

In the current study, lower PWS and tSNR values were found for pCASL compared with FAIR. This finding is not in line with the notion that the inherent SNR is higher for pCASL than for PASL techniques such as FAIR. For the brain, this has been shown by theoretical modeling of the perfusion signal, and has been demonstrated experimentally [9, 38]. The higher intrinsic SNR of pCASL in the brain builds first on the longer temporal duration of the labeled bolus, which is proportional to a larger volume of labeled

blood that is delivered to the tissue and second on the closer proximity of the labeling location and the imaging slab, which reduces T1 decay. The geometry of pCASL used in the kidney, with labeling taking place approximately 15 cm upstream of the tissue, reverses the labeling proximity argu-ment to the advantage of FAIR. For FAIR, spatial coverage of the non-selective labeling slab is limited by the transmit RF coil, resulting in a smaller volume of labeled blood in brain. However, for kidneys, body coil transmission is used and kidneys are positioned nearly at isocenter, so this is expected to result in a very minor difference in the labeled blood volume compared to pCASL. In addition, FAIR has an essentially flow velocity independent inversion efficiency [38], which may be beneficial for application areas with a broad distribution of flow velocities and pulsatile flow.

As previously indicated, the poorer performance of pCASL might also be a result of reduced labeling efficiency caused by (a combination of) several factors. First, labeling efficiency is sensitive to B0 inhomogeneities present at the pCASL labeling location (due to the proximity of air in the lungs), especially for this study performed at 3T. The labe-ling efficiency of the balanced pCASL implementation used is more sensitive to B0 offsets than that of the unbalanced variant. However, separate B0 shimming at the labeling loca-tion during acquisition was performed to mitigate this effect. Still, B0 shimming at the labeling location may fail, thus resulting in compromised labeling efficiency. Second, the presence of B1 inhomogeneities may have resulted in lower B1 than expected at the labeling location. Finally, in contrast to FAIR, the inversion efficiency of pCASL is dependent on flow velocity [38]. In the current study, default implemented labeling settings were used for pCASL, which have been optimized for brain application. Blood flow characteristics such as maximum blood flow velocity and pulsatility of the descending aorta are typically different compared with those in the brain feeding arteries [39, 40]. At higher flow velocities, the adiabatic condition will be violated resulting in less optimal inversion of blood spins [38]. Optimization of pCASL labeling parameters for renal application has been shown to improve robustness to off-resonance effects and aortic flow pulsatility [41]. To improve repeatability of per-fusion quantification and to detect failed measurements on a subject level, we think that it is recommended to measure the labeling efficiency at the labeling location, similar as has been proposed for the brain [42, 43]. This will permit to judge the technical validity of the ASL measurement and to correct for the labeling efficiency in the quantification.

This study has limitations. First, for the pCASL labeling approach, two variants have been proposed, balanced and unbalanced [18]. The balanced variant, which was used in this study, has been shown to be more sensitive to B0 off-sets. Switching to unbalanced pCASL may improve robust-ness to off-resonance effects, as has been demonstrated for

Page 12: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

92 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

brain ASL [44] and preliminary data is available that shows this for the kidneys as well [45]. Nevertheless, unbalanced pCASL has its own disadvantages such as vulnerability to subtraction artifacts caused by eddy currents due to the usage of different gradient waveforms in label and control condi-tion [18]. A more extensive direct comparison between both pCASL variants should be made in future studies to further optimize this labeling approach and improve robustness. Second, renal perfusion lacks a gold standard technique to enable validation of RBF values obtained with both labe-ling approaches. Alternative techniques for measurement of renal perfusion include the para-aminohippurate (PAH) clearance method [7] and PET imaging (using 15O-labeled water) [46]. PAH clearance measurement involves multiple blood and urine samples over a time course of several hours and provides only information on total perfusion of both kidneys combined. PET imaging involves the infusion of a radioactive tracer, but enables localized perfusion measure-ments. However, even in the absence of a gold standard, the clinical value of a perfusion measurement technique is determined by its capability to measure clinically relevant perfusion differences and changes. Finally, the acquisition order of FAIR and pCASL scans was not randomized. All FAIR scans with varying delay times were always performed before the pCASL scans, resulting in a difference of ~ 10 min between acquisition of both labeling approaches. This small time difference will probably not explain the difference found between the two methods.

In conclusion, in this comparative study between multi-delay FAIR and balanced pCASL for renal perfusion measurements at 3T in healthy middle-aged volunteers, FAIR showed favorable repeatability. To improve repeat-ability of perfusion quantification and assess the technical validity of an ASL measurement, addition of a labeling efficiency measurement is recommended, especially for the balanced pCASL variant used in this study.

Acknowledgements We thank MeVis Medical Solutions AG (Bremen, Germany) for providing MeVisLab medical image processing and visu-alization environment, which was used for image processing and analy-sis. This work is part of the research program Applied and Engineering Sciences with project number 14951 which is (partly) financed by the Netherlands Organization for Scientific Research (NWO). A.B. was supported by an Alexandre Suerman stipend for MD-PhD students by the University Medical Center Utrecht, The Netherlands.

Author contributions AAH: study conception and design, acquisition of data, analysis and interpretation of data, drafting of manuscript, and critical revision. ADB: study conception and design, acquisition of data, drafting of manuscript, and critical revision. SLF: study concep-tion and design, and critical revision. TL: study conception and design, and critical revision. MVS: study conception and design, drafting of manuscript, and critical revision. CB: study conception and design, drafting of manuscript, and critical revision.

Compliance with ethical standards

Conflict of interest There are no conflicts of interest related to this work.

Ethical approval All procedures performed in studies involving human participants were in accordance with the ethical standards of the institu-tional research committee [Medical Ethics Advisory Committee Utre-cht (METC Utrecht) reference number NL.45144.041.13] and with the 1964 Helsinki declaration and its later amendments or comparable ethical standards.

Informed consent Informed consent was obtained from all individual participants included in the study.

Open Access This article is distributed under the terms of the Crea-tive Commons Attribution 4.0 International License (http://creat iveco mmons .org/licen ses/by/4.0/), which permits unrestricted use, distribu-tion, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made.

References

1. Selby NM, Blankestijn PJ, Boor P, Combe C, Eckardt KU, Eikef-jord E, Garcia-Fernandez N, Golay X, Gordon I, Grenier N, Hock-ings PD, Jensen JD, Joles JA, Kalra PA, Kramer BK, Mark PB, Mendichovszky IA, Nikolic O, Odudu A, Ong ACM, Ortiz A, Pruijm M, Remuzzi G, Rorvik J, de Seigneux S, Simms RJ, Slatin-ska J, Summers P, Taal MW, Thoeny HC, Vallee JP, Wolf M, Car-oli A, Sourbron S (2018) Magnetic resonance imaging biomarkers for chronic kidney disease: a position paper from the European Cooperation in Science and Technology Action PARENCHIMA. Nephrol Dial Transpl 33(suppl_2):ii4–ii14

2. Odudu A, Nery F, Harteveld AA, Evans RG, Pendse D, Buchanan CE, Francis ST, Fernandez-Seara MA (2018) Arterial spin label-ling MRI to measure renal perfusion: a systematic review and statement paper. Nephrol Dial Transpl 33(suppl_2):ii15–ii21

3. Winter JD, St Lawrence KS, Cheng HL (2011) Quantification of renal perfusion: comparison of arterial spin labeling and dynamic contrast-enhanced MRI. J Magn Reson Imaging 34(3):608–615

4. Sourbron SP, Michaely HJ, Reiser MF, Schoenberg SO (2008) MRI-measurement of perfusion and glomerular filtration in the human kidney with a separable compartment model. Invest Radiol 43(1):40–48

5. Dujardin M, Sourbron S, Luypaert R, Verbeelen D, Stadnik T (2005) Quantification of renal perfusion and function on a voxel-by-voxel basis: a feasibility study. Magn Reson Med 54(4):841–849

6. Ritt M, Janka R, Schneider MP, Martirosian P, Hornegger J, Bautz W, Uder M, Schmieder RE (2010) Measurement of kidney perfu-sion by magnetic resonance imaging: comparison of MRI with arterial spin labeling to para-aminohippuric acid plasma clearance in male subjects with metabolic syndrome. Nephrol Dial Transpl 25(4):1126–1133

7. Cole BR, Giangiacomo J, Ingelfinger JR, Robson AM (1972) Measurement of renal function without urine collection. A critical evaluation of the constant-infusion technic for deter-mination of inulin and para-aminohippurate. N Engl J Med 287(22):1109–1114

Page 13: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

93Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

8. Williams DS, Detre JA, Leigh JS, Koretsky AP (1992) Magnetic resonance imaging of perfusion using spin inversion of arterial water. Proc Natl Acad Sci USA 89(1):212–216

9. Alsop DC, Detre JA, Golay X, Gunther M, Hendrikse J, Her-nandez-Garcia L, Lu H, MacIntosh BJ, Parkes LM, Smits M, van Osch MJ, Wang DJ, Wong EC, Zaharchuk G (2015) Recom-mended implementation of arterial spin-labeled perfusion MRI for clinical applications: a consensus of the ISMRM perfusion study group and the European consortium for ASL in dementia. Magn Reson Med 73(1):102–116

10. Wong EC (2014) An introduction to ASL labeling techniques. J Magn Reson Imaging 40(1):1–10

11. Levey AS, Stevens LA, Schmid CH, Zhang YL, Castro AF 3rd, Feldman HI, Kusek JW, Eggers P, Van Lente F, Greene T, Coresh J (2009) A new equation to estimate glumerular filtration rate. Ann Intern Med 150(9):604–612

12. Martirosian P, Klose U, Mader I, Schick F (2004) FAIR true-FISP perfusion imaging of the kidneys. Magn Reson Med 51(2):353–361

13. Ordidge RJ, Wylezinska M, Hugg JW, Butterworth E, Franconi F (1996) Frequency offset corrected inversion (FOCI) pulses for use in localized spectroscopy. Magn Reson Med 36(4):562–566

14. Yongbi MN, Branch CA, Helpern JA (1998) Perfusion imaging using FOCI RF pulses. Magn Reson Med 40(6):938–943

15. Ogg RJ, Kingsley PB, Taylor JS (1994) WET, a T1- and B1-insen-sitive water-suppression method for in vivo localized 1H NMR spectroscopy. J Magn Reson B 104(1):1–10

16. Wong EC, Buxton RB, Frank LR (1998) Quantitative imaging of perfusion using a single subtraction (QUIPSS and QUIPSS II). Magn Reson Med 39(5):702–708

17. Li X, Auerbach EJ, Van de Moortele PF, Ugurbil K, Metzger GJ (2018) Quantitative single breath-hold renal arterial spin labeling imaging at 7T. Magn Reson Med 79(2):815–825

18. Wu WC, Fernandez-Seara M, Detre JA, Wehrli FW, Wang J (2007) A theoretical and experimental investigation of the tag-ging efficiency of pseudocontinuous arterial spin labeling. Magn Reson Med 58(5):1020–1027

19. Clare S, Jezzard P (2001) Rapid T(1) mapping using multislice echo planar imaging. Magn Reson Med 45(4):630–634

20. Schalkx HJ, Petersen ET, Peters NH, Veldhuis WB, van Leeuwen MS, Pluim JP, van den Bosch MA, van Stralen M (2015) Arterial and portal venous liver perfusion using selective spin labelling MRI. Eur Radiol 25(6):1529–1540

21. Huizinga W, Poot DH, Guyader JM, Klaassen R, Coolen BF, van Kranenburg M, van Geuns RJ, Uitterdijk A, Polfliet M, Vande-meulebroucke J, Leemans A, Niessen WJ, Klein S (2016) PCA-based groupwise image registration for quantitative MRI. Med Image Anal 29:65–78

22. Klein S, Staring M, Murphy K, Viergever MA, Pluim JP (2010) elastix: a toolbox for intensity-based medical image registration. IEEE Trans Med Imaging 29(1):196–205

23. Buxton RB, Frank LR, Wong EC, Siewert B, Warach S, Edel-man RR (1998) A general kinetic model for quantitative per-fusion imaging with arterial spin labeling. Magn Reson Med 40(3):383–396

24. Zhang X, Petersen ET, Ghariq E, De Vis JB, Webb AG, Teeuwisse WM, Hendrikse J, van Osch MJ (2013) In vivo blood T(1) meas-urements at 1.5T, 3T, and 7T. Magn Reson Med 70(4):1082–1086

25. Herscovitch P, Raichle ME (1985) What is the correct value for the brain–blood partition coefficient for water? J Cereb Blood Flow Metab 5(1):65–69

26. R Core Team (2014) R: a language and environment for statistical computing. https ://www.r-proje ct.org

27. Nery F, Gordon I, Thomas DL (2018) Non-invasive renal perfu-sion imaging using arterial spin labeling MRI: challenges and

opportunities. Diagnostics (Basel). https ://doi.org/10.3390/diagn ostic s8010 002

28. Buchanan CE, Cox EF, Francis ST (2018) Evaluation of 2D imag-ing schemes for pulsed arterial spin labeling of the human kidney cortex. Diagnostics (Basel). https ://doi.org/10.3390/diagn ostic s8030 043

29. Moore JE Jr, Ku DN (1994) Pulsatile velocity measurements in a model of the human abdominal aorta under simulated exercise and postprandial conditions. J Biomech Eng 116(1):107–111

30. Taylor CA, Cheng CP, Espinosa LA, Tang BT, Parker D, Herfkens RJ (2002) In vivo quantification of blood flow and wall shear stress in the human abdominal aorta during lower limb exercise. Ann Biomed Eng 30(3):402–408

31. Gillis KA, McComb C, Patel RK, Stevens KK, Schneider MP, Radjenovic A, Morris ST, Roditi GH, Delles C, Mark PB (2016) Non-contrast renal magnetic resonance imaging to assess perfu-sion and corticomedullary differentiation in health and chronic kidney disease. Nephron 133(3):183–192

32. Prasad PV, Li LP, Thacker JM, Li W, Hack B, Kohn O, Sprague SM (2019) Cortical perfusion and tubular function as evaluated by magnetic resonance imaging correlates with annual loss in renal function in moderate chronic kidney disease. Am J Nephrol 49(2):114–124

33. Cox EF, Buchanan CE, Bradley CR, Prestwich B, Mahmoud H, Taal M, Selby NM, Francis ST (2017) Multiparametric renal mag-netic resonance imaging: validation, interventions, and alterations in chronic kidney disease. Front Physiol 8:696

34. Gillis KA, McComb C, Foster JE, Taylor AH, Patel RK, Morris ST, Jardine AG, Schneider MP, Roditi GH, Delles C, Mark PB (2014) Inter-study reproducibility of arterial spin labelling mag-netic resonance imaging for measurement of renal perfusion in healthy volunteers at 3 Tesla. BMC Nephrol 15:23

35. Kim DW, Shim WH, Yoon SK, Oh JY, Kim JK, Jung H, Matsuda T, Kim D (2017) Measurement of arterial transit time and renal blood flow using pseudocontinuous ASL MRI with multiple post-labeling delays: feasibility, reproducibility, and variation. J Magn Reson Imaging 46(3):813–819

36. Conlin CC, Oesingmann N, Bolster B Jr, Huang Y, Lee VS, Zhang JL (2017) Renal plasma flow (RPF) measured with multiple-inver-sion-time arterial spin labeling (ASL) and tracer kinetic analysis: validation against a dynamic contrast-enhancement method. Magn Reson Imaging 37:51–55

37. Heyman SN, Khamaisi M, Rosen S, Rosenberger C (2008) Renal parenchymal hypoxia, hypoxia response and the progression of chronic kidney disease. Am J Nephrol 28(6):998–1006

38. Wong EC, Buxton RB, Frank LR (1998) A theoretical and experi-mental comparison of continuous and pulsed arterial spin labeling techniques for quantitative perfusion imaging. Magn Reson Med 40(3):348–355

39. Amanuma M, Mohiaddin RH, Hasewaga M, Heshiki A, Longmore DB (1992) Abdominal aorta: characterisation of blood flow and measurement of its regional distribution by cine magnetic reso-nance phase-shift velocity mapping. Eur Radiol 2:559–564

40. Oglat AA, Matjafri MZ, Suardi N, Oqlat MA, Abdelrahman MA, Oqlat AA (2018) A review of medical doppler ultrasonography of blood flow in general and especially in common carotid artery. J Med Ultrasound 26(1):3–13

41. Echeverria-Chasco R, Vidorreta M, Aramendia-Vidaurreta V, Bas-tarrika G, Fernandez-Seara MA (2019) Optimization of pseudo continuous arterial spin labeling for renal ASL. In: Proceedings of the 27th scientific meeting, International Society for Magnetic Resonance in medicine, Montreal, Canada:#4954

42. Chen Z, Zhang X, Yuan C, Zhao X, van Osch MJP (2017) Meas-uring the labeling efficiency of pseudocontinuous arterial spin labeling. Magn Reson Med 77(5):1841–1852

Page 14: C‑delayFAIRCASLoaches for˜renal perfusion quanti˚cation …82 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94 1 3 canachieveahigherintrinsicsignal-to-noiseratio(SNR)

94 Magnetic Resonance Materials in Physics, Biology and Medicine (2020) 33:81–94

1 3

43. Chen Z, Zhao X, Zhang X, Guo R, Teeuwisse WM, Zhang B, Koken P, Smink J, Yuan C, van Osch MJP (2018) Simultaneous measurement of brain perfusion and labeling efficiency in a single pseudo-continuous arterial spin labeling scan. Magn Reson Med 79(4):1922–1930

44. Zhao L, Vidorreta M, Soman S, Detre JA, Alsop DC (2017) Improving the robustness of pseudo-continuous arterial spin labe-ling to off-resonance and pulsatile flow velocity. Magn Reson Med 78(4):1342–1351

45. Greer J, Wang Y, Pedrosa I, Madhuranthakam A (2019) Pseudo-continuous arterial spin labeled renal perfusion imaging at 3T with improved robustness to off-resonance. In: Proceedings of

the 27th scientific meeting, International Society for Magnetic Resonance in medicine, Montreal, Canada:#4959

46. Green MA, Hutchins GD (2011) Positron emission tomog-raphy (PET) assessment of renal perfusion. Semin Nephrol 31(3):291–299

Publisher’s Note Springer Nature remains neutral with regard to jurisdictional claims in published maps and institutional affiliations.